Hi @Steven_Hines,
32 minutes ? That’s awesome.
Can you take apart your dispenser and take a look at your ink cartridge? It is possible the piston seal in that cartridge has broken. A little bit of ink crawling up the piston edges is normal, but excessive amounts of ink is not. From your description I do not think this is happening but just in case.
One parameter that you can adjust is the Rheological setpoint. We explain what each parameter does here. But long story short - increasing this parameter will increase the ink flow over time.
This parameter is quite sensitive, in the example below I’ve increased it to 0.20 for the ExcitableSnail profile. Can you give this a try? If you still have issues - try 0.21 or 0.22.
One important thing:
The V-One can detect when the probe is ON or OFF but it cannot detect when the dispenser is mounted/removed. So if the system has performed the switch calibration and you then physically remove the dispenser, it is important to force the system to re-calibrate. If it does not re-calibrate, your height will be off and you risk broken nozzles.
Clicking the ‘Home’ icon or pressing Alt+R to release the motors will force the system to re-calibrate the next time you click ‘Start’